I’ve run a salt pool for two years. My Eagle Ray is great for pH & FC, but I treat its salt number as a ‘compass,’ not gospel. It reads salinity via conductivity and temperature can swing it by ~100–200 ppm. What helps: a monthly calibration with 0 ppm (RO) and 3000 ppm solution, let the cuvette/probe sit in pool water for ~2 minutes to match temperature, and pull the sample elbow-deep away from returns/the cell. For a true number I use a drop titration (e.g., K-1766) or compare with the salt cell’s panel; if those differ by >300 ppm, I trust the titration. I aim for 3200–3400 ppm to leave a rainy-season buffer. One more tip: when adding salt, split the dose (⅓–½), dissolve around the pool, run the pump 24 hours, then retest.
By the way, which salt cell brand are you using? Different units suggest different target ppm.